EPAG Perspective - Regulatory Advances Related to Nasal Spray Pumps. Dr G.Williams Nasal Drug Delivery Management Forum London, 15 Apr 2010

Similar documents
The Role of EPAG in Standards and Regulatory Guidance Development

Understanding Regulatory Global Requirements for Nasal Drug Products. Julie D. Suman, Ph.D. April 8, 2016

Inhalation Product Research at FDA

Caption: The equipment required for testing Fluticasone Propionate (FP) Inhalation Powder in line with a new product-specific monograph (USP36-NF31).

COMMENTS. Submitted by The International Pharmaceutical Aerosol Consortium

COMMITTEE FOR MEDICINAL PRODUCTS FOR HUMAN USE (CHMP) GUIDELINE ON THE PHARMACEUTICAL QUALITY OF INHALATION AND NASAL PRODUCTS

OPTIMISING ANALYTICAL STRATEGIES FOR THE DEMONSTRATION OF BIOEQUIVALENCE IN A GENERIC NEBULISER

Drug/Device Combination Products: Bioequivalence

Understanding cascade impaction and its importance for inhaler testing

PRACTICES OF COATING COLLECTION SURFACES OF CASCADE IMPACTORS: A SURVEY OF MEMBERS OF THE EUROPEAN PHARMACEUTICAL AEROSOL GROUP (EPAG)

Assessing Quality of Inhaled Products And Links to Efficacy and Safety

Adapting Pharmacopeial Testing for Pharmaceutical Equivalence Studies. Julie D. Suman, Ph.D. March 29, 2016

DISSOLUTION TESTING - Exploring the Link Between Particle Size & Dissolution Behavior for OINDPs

Implications of the new CHMP Guideline on the Pharmaceutical Quality of Inhalation and Nasal Products

Go With the Flow REGULATORY LANDSCAPE. Mark Copley at Copley Scientific

Assessing the role of breathing simulators in OIP testing

Effects of viscosity, pump mechanism and nozzle geometry on nasal spray droplet size

Qualifying Container Closure Systems for OINDP: Current & Future Regulatory Expectations. Julie D. Suman, Ph.D. November 14, 2014

Device Change Management for Inhaled Products. Loy Britto, Ph.D. GlaxoSmithKline ISAM Congress Munich 2015

Equivalence Evaluation of Valved Holding Chambers (VHCs) with Albuterol Pressurized Metered Dose Inhaler (pmdi)

Testing Inhaled Generics

What Do Cascade Impaction Measurements Tell Us: In Vitro Aspects. Context of Presentation

International Pharmaceutical Aerosol Consortium on Regulation and Science

INTRODUCTION. size and total nozzle area decrease with stage number. Volumetric air flow rate through

EPAG Sponsored Workshop on Abbreviated Impactor Measurement (AIM) and Efficient Data Analysis (EDA) Concepts in Inhaler Testing. Overview of AIM EDA

Quality of Nasal and Inhalation Drug Products

OVERVIEW OF EFFECTIVE DATA ANALYSIS (EDA)

Lessons Learned from Approval of Generic Nasal Products

Metered Dose Inhalers (MDIs) In Vitro Measures to Confirm Patient Perceptions: HFA vs. CFC

Process Drift and it s Resolution in the Manufacture of Drug Products. MDI s and DPI s Metered Dose Inhalations and Dry Powder Inhalations

Brief Overview of the PQRI Recommendations: Challenges and Successes

Current Challenges and Opportunities in Demonstrating Bioequivalence

Technology and Quality of Inhalation Drug Products

Considerations for Evaluation of Bioequivalence and Interchangeability of Orally Inhaled Products

REDUCING VARIABILITY IN TEST RESULTS FOR OINDPS WITH AUTOMATED ACTUATION

Significance of Leachables and Extractables to Pharmaceutical Quality

Public Assessment Report. Scientific discussion. Elisox 50 microgram/actuation, nasal spray, suspension. (mometasone furoate) NL/H/2672/001/DC

Best Practices for OINDP Pharmaceutical Development Programs Leachables and Extractables

Last updated: July 2011

The abbreviated impactor measurement concept

905 UNIFORMITY OF DOSAGE UNITS

University of Groningen. Technology in practice Lexmond, Anne

Guidance for Industry

DEBATING THE OPERATING CURVES FOR DDU TESTS ON MARKETED INHALERS

UPDATE ON THE GUIDELINE ON THE PHARMACEUTICAL QUALITY OF INHALATION AND NASAL PRODUCTS

Characterizing the Performance of Metered Dose Inhalers with Add-On Devices: New Methods For Clinically Relevant Testing

Challenges in Nonclinical Development of Inhalation Drug Products

Metered Dose Inhaler Technology

Product Guide. MSP Corporation Rice Creek Parkway, Suite 300. Shoreview, Minnesota 55126, U.S.A. Phone: Fax:

COPLEY SCIENTIFIC. Quality Solutions for Inhaler Testing. metered-dose inhalers dry powder inhalers nebulisers nasal sprays

Meeting the Quality Challenge for Orally Inhaled Drug Products. Review Impact of Emerging EMA/MHRA Guidelines and Standards

Nonclinical Safety Evaluation of Inhalation Drug Products

Everything for Inhalation

General Chapter/Section: <232> Elemental Impurities - Limits Expert Committee(s): General Chapters Chemical Analysis No.

Use of Math Modelling to Understand Delivery of Biopharmaceutical Molecules to the Lung

Year in review. Vit Perlik Director of Regulatory Science and Clinical Development

BCS: Dissolution Testing as a Surrogate for BE Studies

Case Study 1: Pharmaceutical Development of EXUBERA

IVIVC in Pediatric OIPs

Generic Inhaled Medications

ADVANCIA : A TRUE PLATFORM & INNOVATIVE SYSTEM OFFERING EFFICIENT NASAL DOSE DELIVERY

Biopharmaceutics of Non-Orally Administrated Drugs

Dry Powder Inhaler. Developing an Efficient. 3M Conix DPI. White Paper / Spring Proven Solutions that Enable Your Success

SELECTION AND VALIDATION OF CASCADE IMPACTOR TEST METHODS

SPS Pharma: Who we are?

PREMIXES FOR MEDICATED FEEDING STUFFS FOR VETERINARY USE. Praeadmixta ad alimenta medicata ad usum veterinarium. Effervescent powders

Draft Guidance on Fluticasone Propionate; Salmeterol Xinafoate. Fluticasone Propionate; Salmeterol Xinafoate. Powder/Inhalation

IPAC-RS/UF Orlando Inhalation Conference March 20, S.T. Horhota 1, C.B. Verkleij 2, P.J.G. Cornelissen 2, L. Bour 3, A. Sharma 3, M.

Citation for published version (APA): Westerman, E. M. (2009). Studies on antibiotic aerosols for inhalation in cystic fibrosis s.n.

COMPARISON OF THE RESPIRABLE FRACTION FROM THREE DIFERENT DPI DEVICES

Device Design Similarity

COPLEY SCIENTIFIC. Quality Solutions for Inhaler Testing Edition. metered-dose inhalers dry powder inhalers nebulisers nasal sprays

Misty Max 10 nebulizer

ISO Process and Standards Under Development. ISO Principles

Respiratory. Martin Jetzer DDL27 Edinburgh December 2016

Ensuring the Quality of Some Drug-Device Combination Products - FDA Perspective

AAM Fall Tech Conference November 6-8, 2017 Rockville, MD

In-Vitro Bioequivalence Studies for Oral Solid Dose products using the Morphologi G3-ID

RDD Europe 2011 Workshop 4 May 2011

AEROSOL THERAPY: THE PRACTICALITIES

IMPROVING THE REALISM AND RELEVANCE OF MOUTH-THROAT MODELS FOR INHALED PRODUCT TESTING

Work plan for the joint CHMP/ CVMP Quality Working Party (QWP) for 2018

VARIABILITY IN CASCADE IMPACTION: SOURCES, IMPACT AND STRATEGIES FOR REDUCTION

<1151> Pharmaceutical Dosage Forms and <1152> Animal Drugs for Use in Animal Feeds

Bioequivalence Requirements: USA and EU

DRUG PRODUCT PERFORMANCE: CONSIDERATIONS FOR INTERCHANGEABILITY OF MULTISOURCE DRUG SUBSTANCES AND DRUG PRODUCTS

Completion of the development of a formulation: Requirements for compliance check vs. requirements for Marketing Authorisation

THE DEVELOPMENT PATH TO BIOEQUIVALENCE FOR A NASAL SPRAY

Performance of a Fluticasone Propionate/Salmeterol Xinafoate 3M Taper DPI

Should you have any questions, please contact Ravi Ravichandran, Principal Scientific Liaison ( or

ANDA Submissions Refuse to Receive for Lack of Proper Justification of Impurity Limits Guidance for Industry

Metered Dose Inhalation Formulations of Salbutamol Sulphate Using Non-CFC Propellant Tetrafluoroethane

6/7/2017 CHANGES IN PHARMA REGULATIONS HOW IT AFFECTS THE OR NURSE DISCLOSURE OBJECTIVES

Partner with the Global Leader in Drug Delivery Systems.

Partner with the Global Leader in Drug Delivery Systems.

University of Groningen. Optimisation of dry powder inhalation Boer, Anne Haaije de

Dr. Christian Zeine LGC Standards GmbH. Webinar Series 2013 July 2013

Optimising the application of in vitro test methods for the demonstration of bioequivalence in orally inhaled products

INTEGRATED DESIGN SPACE TO DEVELOP BETTER DPI FORMULATIONS

Intranasal Drug Administration An Attractive Delivery Route for Some Drugs

Transcription:

1 EPAG Perspective - Regulatory Advances Related to Nasal Spray Pumps Dr G.Williams Nasal Drug Delivery Management Forum London, 15 Apr 2010

Overview 2 EPAG, what is it?, background, this project Regulatory guidance Pharmacopeial guidance What current in-vitro testing methods can and can not tell us Defining mass fraction penetrating the nasopharynx Methods applied and results to date for % particles <10µm Next steps

EPAG, what is it?, background, this project 3 EPAG Organisation (European Pharmaceutical Aerosol Group) Voluntary organisation Self managed Annual fee (non-profit making basis), Funding studies, experts etc Rules & Constitution EPAG Membership, current Feb 2010 3M, Almirall Sofotec, AstraZeneca, Bespak Europe Ltd, Boehringer Ingelheim, Chiesi, Clinical Designs Ltd, GlaxoSmithKline, Hovione, Novartis, Pari, Pfizer, sanofi-aventis, Siegfried Pharma Development, SkyePharma, Teva, Trudell Medical International, Valois, Vectura

EPAG, what is it?, background, this project 4 Objectives Focus on pharmaceutical issues relevant to Pulmonary and Nasal Delivery products, including clinical aspects as appropriate Establish scientifically based best practices Provide consensus comment to Industry and Government Agencies to promote safety and quality standards Recommend harmonised standards and methodology

EPAG, what is it?, background, this project 5 EPAG Membership Membership is open to: European Pharmaceutical Companies that develop new drug product for human use utilising the Pulmonary or Nasal route of delivery Product development takes place in collaboration with other Pharmaceutical companies as well as non-pharmaceutical companies (e.g. API suppliers, device design and componentry manufacturers, consultants etc). These companies may be invited to attend when a particular need is identified. Companies, including non-european companies, may be admitted as members where they provide specific expertise that supports and enhances EPAG objectives

EPAG, what is it?, background, this project 6 This project, European pharmacopeia methodology for Nasal sprays Aerodynamic particle size analysis of two fractions. (< 10µm >) Methodology produces data as representative as possible of the spray exiting the device orifice Method validation package

REGULATORY: Heath Canada-EMEA 7 Heath Canada-EMEA: Joint Guideline on the Pharmaceutical Quality of Inhalation and Nasal Products EMEA/CHMP/QWP/49313/2005 Corr - Section 4.2.2 The tests indicated in Table 4.2.2 are normally conducted to characterize nasal products Not all tests are necessary for all types of nasal products (see guideline for details)

REGULATORY: Heath Canada-EMEA 8 Covers 6 distinct categories of nasal delivery device Pressurized metered dose nasal sprays Nasal powders (device metered) Nasal liquids Single use drops Multiple use drops Single use sprays Non pressurized multiple use metered dose sprays

REGULATORY: Heath Canada-EMEA 9 Pharmaceutical development studies Physical characterization Minimum fill justification Extractables/leachables Delivered dose uniformity through container life Particle/droplet size distribution Actuator deposition Shaking requirements Initial & re-priming requirements Cleaning requirements Low temperature performance Performance after temperature cycling Effect of environmental moisture robustness Delivery device development

US Regulatory requirements 10 USFDA (CDER): July 2002: Guidance for Industry - Nasal Spray and Inhalation Solution, Suspension, and Spray Drug Products CMC Documentation April 2003: Draft Guidance for Industry - Bioavailability and Bioequivalence Studies for Nasal Aerosols and Nasal Sprays for Local Action

CMC documentation: Nasal sprays 11 Description Identification Assay Impurities and Degradation Products Spray Content Uniformity Preservatives and Pump Delivery Stabilizing Excipients Assay Spray Pattern and Plume Geometry Pump Delivery Droplet Size Distribution Particle Size Distribution (Suspensions) Particulate Matter Microbial Limits Net Content Weight Loss (Stability) Leachables (Stability) ph Osmolality Viscosity

Inhalation solutions, suspensions and sprays 12 Description Identification Assay Impurities and Degradation Products Preservatives and stabilizing Excipients Assay Sterility Particulate matter ph Osmolality Net content Weight Loss (Stability) Leachables (Stability) Particle size distribution (suspensions) Pump delivery for inhalation sprays Spray content uniformity for inhalationsprays Plum geometry for inhalation sprays Particle/droplet size distribution for inhalation sprays -

Draft Nasal BA/BE guidance 13 Single Actuation Content Through Container Life Drug in Small Particles/Droplets, or Particle/Droplet Size Distribution by Cascade Impactor Spray pattern Droplet Size Distribution by Laser Diffraction Drug Particle Size Distribution by Microscopy Plume geometry Priming and re-priming -

Pharmacopeial guidance 14 Ph. Eur. 6th Edition (2010): Nasal Preparations Nasalia : Tests for uniformity of dosage units, mass, content for nasal drops/liquid nasal sprays Monograph undergoing revision to include droplet/particle sizing methodologies as well as more detail on mean delivered dose and dose uniformity Inhalanda Committee USP: USP 32/NF 27 (2009): Nasal Sprays are included as one of the oral and nasal inhaled drug products (OINDPs) under <601> Physical Tests & Determinations Tests for delivered dose uniformity N.B. <601> includes methods for particle sizing developed for oral rather than nasal inhaled preparations

What current in-vitro methods can and can not tell us 15 CAN PROVIDE MEASURES OF: Dose content Dose content uniformity through life Entire droplet and particle size distribution not traceable to drug substance by currently accepted method of laser diffractometry Spray pattern and geometry LIMITATIONS: No standardized entry port mimicking nares and nasal vestibule Necessary if mass fractions are required Mass fraction < ca. 10 μm aerodynamic diameter likely to penetrate beyond the nasopharynx New Test Needed

Defining the mass fraction penetrating the nasopharynx 16 1st STEP: Define standardized entry port: Doub and Adams (2002 AAPS Annual Meeting) explored various sizes of inverted round bottomed flasks for use testing nasal spray pumps Linked with abbreviated cascade impactor to determine mass fraction <10 μm Sample at 28.3 L/min courtesy W. Doub, FDA, St. Louis, MO

Define standard entry port 17 Doub and Adams (2002): 1 L flask was too small Larger (2 L and 5 L flasks) required a pre-separator Observations-1: Choice of a fixed flow rate was limiting: Adult: 10-40 L/min Häußermann et al. J. Aerosol Sci. (2002);33:929-933. Adult: 7 50 L/min Child: 5 20 L/min Cheng AS&T. (2003); 37:659-671 Observations-2: Even a 2 L flask is much larger than the ca. 15 ml volume of the nasal cavity courtesy J.D. Suman, Nextbreath LLC

Define standard entry port 18 JM Aiache (2002): Glass nasal induction port For use with twin impinger Observations: Recoveries between 98-107% Used at 30 & 60lpm Used commercially for a number of years courtesy J.M.Aiache

Nasal induction port design 19 Work is in progress within the Nasal Delivery Sub- Team of the European Pharmaceutical Aerosol Group (EPAG) to develop an induction port that more closely approximates in internal volume and orientation presented to the inhaler as would be the case when used by a patient Possible nasal induction port configurations courtesy Copley Scientific Ltd, and 3M DDSD

Defining the mass fraction penetrating the nasopharynx 20 2nd Step: Establish a size-selective cutpoint close to 10 μm aerodynamic diameter: Requires inertial classifier following inlet Impactor pre-separator almost certainly needed if a low volume induction port is used 3rd Step: Optimize choice of inertial impactor: A conventional 7 or 8 stage full resolution system could be used However, abbreviated 1 or 2 stage impactors are becoming available: Greatly simplify the measurement process More rapid measurement possible better coverage/decision making for batch disposition in QC

Size, characterizing the entire dose 21 Inertial impaction methods are ineffective for droplets > 20 μm aerodynamic diameter Mass median droplet sizes for nasal spray pumps are typically in the 40-80 μm range Laser Diffractometry (LD) has become the industry standard Recommended technique for nasal sprays and nasal aerosols in the FDA BA/BE Draft Guidance Malvern Spraytec LD with nasal spray pump and automated actuator Courtesy Malvern Instruments

Time averaged v-psd 22 Indicative of overall inhaler performance during selected sampling period Less useful than time dependent mode for studying inhaler operation through a complete actuation cycle Volume (%) 20 100 90 80 70 60 10 50 40 30 20 10 0 0 0.1 1.0 10.0 100.0 1000.0 Diametre des Particules (µm.)

Time averaged v-psd 23 Capable of viewing transient phenomena of millisecond duration: Spray/aerosol formation Spray/aerosol development during operation of inhaler Start up stable phase tail off

Developing technology with API specificity 24 LD, though versatile, does not provide chemical species identification linked to the droplet PSD: A problem for multicomponent formulations or where non-active excipient(s) are present or even with single component, suspension APIs

Raman chemical imaging 25 Raman shifts specific to API(s) excipients etc. enable chemical composition of particles imaged to be displayed in a quantitative manner Optical image API identified by colour courtesy J.D. Suman, Nextbreath LLC

Next steps for API specific imaging 26 Extend RCI method to identify additional formulation products Develop automated analytical methods to enable more representative sampling Develop algorithms fusing brightfield microscopy data and RCI to improve particle sizing Compare RCI with standard ingredient-specific methods used to establish bioequivalency between brand and generic nasal products Raman shifts specific to API(s) excipients etc. enable chemical composition of particles imaged to be displayed in a quantitative manner

Challenges. 27 Identify suitable nasal induction port Identify suitable analytical technique and methods for measuring % particles <10µm Method must be robust, validated and acceptable for pharmacopeial inclusion Introduce a standardized method for quantifying % particles <10µm from nasal aerosol drug products..

Methods applied to date. 28 Nasal inlet ports + ACI (short) Inverted NGI + 6L inlet chamber courtesy GSK Inverted NGI + pre-separator courtesy RIVM Laser light scattering

Results to date. 29 Inverted NGI & pre-separator (BDP) None < 10 µm Inverted NGI & 6L chamber (BDP) ~1-2% < 10 µm Nasal induction port & reduced stack ACI (BDP) None < 10µm Malvern Spraytec Laser light scattering (BDP) ~ 0.01% < 10µm Inverted NGI & pre-separator (Xylomzn) <1% < 10 µm Malvern Spraytec Laser light scattering (Xylomzn) ~ 0.3% < 10 µm Marketed Products tested: BDP Nasal spray, Oxymetazoline nasal spray

Next steps. 30 Build data base of %<10µm in nasal sprays by two methods Nasal induction port & cascade impactor Laser light scattering Correlate data, if possible Validate suitable method for EP proposal Make proposal to EP via Inhalanda group Outcome = EP compendial method for assuring levels of fines (<10µm) in nasal sprays to avoid pass-through into the lungs Inclusion in other regulatory standards/pharmacopeias worldwide

Acknowledgments and thanks to. 31 EPAG nasal sub team S.Nichols, J.Mitchell (Trudel), C.Blatchford (3M), K.Hawkins (Teva), JM Aiache (Ind), A.Slater (GSK), B.Doub (FDA), B.Wyka, J.Suman (Nextbreath), T.Purewal, J.Schiewe (BI), M.Weida (RIVM, The Netherlands), G.Williams (Valois, sub-team leader),